I have 3 probes in-line in my HLT and I have a place to test with a certified temp probe, 2 of the probes are within a .1-2 but the other is 2.5-9 off, is there a way other than replacing the probe to compensate for this difference?
Not sure if there is a calculator for Coefficient A, Coefficient B, Coefficient C to adjust for this.
Jerry
Temp Probe - Adjust
Re: Temp Probe - Adjust
There is. One of our forum users here even built a tool to help you set the correct coefficients if you're using the 4.0 beta firmware.
http://bbrally.from-ca.com/
http://bbrally.from-ca.com/